The thing is, ALL tablets have a lot of binding/filler material, so a chromatography assay of a supplement would have way more than just the supplement itself
Also, multiple compounds can have the same molecular weight. It doesn’t determine whether or not you have the right enantiomer…
==
==
like here are lab testing results for semaglutide:
A_User
#2
I don’t care about that or microplastics etc since it is a Pascal’s mugging.
WJ_PhD
#3
In short, no. At least not LCMS as compounds have different ionization potential and so give different detector response for the same amount of material.
Usually purity is checked by LC-UV at a nonspecific wavelength like 220nM. That will at least work for all organic compounds.